Andrew November 10, 2015 at 10:51 am in reply to: BB60c + TG44 in SNA mode resets Store Tru calibration every 2 minutes! //php bbp_reply_id(); ?> Hi Ivan,
There is a new release today. You will see in the release notes, the release yesterday adds the feature that the software will not re-cal when temperature drifts when making scalar measurements. Please note, that accuracy will drift several dB as the device is warming up.
Regards

AndrewModeratorHi Andy,
Working on getting the software to remember the size currently. Expect an update today.
Regards,
A.J.
AndrewModerator
Andrew November 9, 2015 at 3:29 pm in reply to: bb_api.dll x64 linker errors? //php bbp_reply_id(); ?> Hi Cyprien,
If you have installed the Spike x64 application on your PC, you can get the 64-bit version of the ftd2xx.dll library from the Spike installation directory at C:/Program Files/Signal Hound/Spike/
You will see the ftd2xx.dll file that Spike uses, which you can copy into the same directory as the bb_api.dll file for your application. This should resolve the error you are seeing.Regards,
A.J.
AndrewModerator
Andrew November 9, 2015 at 10:07 am in reply to: BB60c + TG44 in SNA mode resets Store Tru calibration every 2 minutes! //php bbp_reply_id(); ?> Hi Ivan,
I have some comments on your feedback. The BB60C will rise to operating temperature when the device is connected to the PC via USB, even if the software is not running. If it is possible, you can connect the unit a bit before you plan to use it, allowing the unit to reach full operating temperature. This will eliminate the annoying re-calibrations. Although, I do agree that it would be better to allow the user to re-calibrate it manually.
I believe you can change the div setting without losing the re-calibration currently. I will determine if this is possible with the reference setting as well in the future.
I agree that with more development, calibrations could be handled in a more consistent manner. If you are able to warm up the unit before your measurements I hope this will eliminate most of the issues you have with this.
Thank you for your feedback as always.
Regards,
A.J.
AndrewModerator
Andrew November 9, 2015 at 9:47 am in reply to: SA44B – external frequency reference & API problem //php bbp_reply_id(); ?> Hi Martin,
Thank you for the heads up on this. This is a known limitation right now. I will be releasing a version today 11/9/2015 that addresses this. The function call will change from saEnableExternalReference to saSetTimebase(device, SA_REF_EXTERNAL_IN). The external timebase will be reset when calling the saCloseDevice function so it should be able to be set again once you re-open the device, without reloading the DLL.
Keep an eye out for that. Version 3.0.9 of the SA api. It will be released with version 3.0.17 of Spike.
Regards,
A.J.
AndrewModerator
Andrew November 9, 2015 at 9:18 am in reply to: Spike on windows 10 x64 remote desktop //php bbp_reply_id(); ?> Hi Cyprien,
We are aware of this issue. The reason this happens is because we utilize the GPU in our application through the OpenGL library. This is not supported in Windows Remote Desktop unfortunately and is a known limitation. There are a few workarounds.
Some customers report that if they can launch Spike before logging in to remote desktop and it will work, but they cannot launch Spike through remote desktop. You can also use a VNC client/server. There is a free version called TightVNC which we have used with success. This lets you control the PC remotely. If you have the NUC with vPro functionality, there is a way to setup the device to be controlled remotely using vPro. This is more involved but is a very robust solution.
Regards,
A.J.
AndrewModerator
Andrew November 9, 2015 at 9:12 am in reply to: using the API to query real-time to get peak to mean ratio CDF etc //php bbp_reply_id(); ?> Hello Cyprien,
The frameWidth is the sweep size divided by an even integer if the sweepSize is >= 1024. This means, the frameWidth will always be between 512 and 1023. Example, If the sweepSize is 1026 then the frameWidth will be 513, and each column in the frame will be the contribution of two bins of the sweep. We always use an even integer divisor so you can consistently determine which sweep bins fall into the frame columns.
The frame returned is the accumulation and scaling of the last 20 frames. It is implemented as a 20 frame FIFO. If you specify the default 30FPS, then each frame is the accumulation of the last 2/3 second.
If you have any further questions, let me know.
Regards,
A.J.
AndrewModeratorIvan,
The overload could be energy in the entire acquisition band you are currently in. In your case, at 10MHz, the device is using the direct RF path which would account for frequency from 9kHz -> 27MHz. If you increase your span from 9kHz to 27MHz do you still see the overload?
Regards,
A.J.
AndrewModerator
Andrew November 5, 2015 at 9:54 am in reply to: BB60c with 10MHz external reference from TG44 shows me large frequency deviation //php bbp_reply_id(); ?> Hi Ivan,
After further discussion here, we are going to add the ability to specify whether you are using the TG 10MHz reference output, so it will disable this frequency correction. We think allowing the user to manually specify this is the best way forward. It will likely be a new checkbox in the TG control panel. Look for this in the next release.
Let me know if you have any further feedback or questions on the issue.
Regards,
A.J.
AndrewModerator
Andrew November 5, 2015 at 9:30 am in reply to: BB60c with 10MHz external reference from TG44 shows me large frequency deviation //php bbp_reply_id(); ?> Hi Ivan,
You are correct, there is a frequency correction offset that is applied to the output of the TG. In the original 2.18 software, the correction value was ignored when the SA is configured to accept an external reference. The Spike software currently does not perform this action. I will look into adding this as a parameter in the software.
Regards,
A.J.
AndrewModerator
Andrew November 5, 2015 at 9:22 am in reply to: BB60c + TG44 – strange jumps on plot in SNA mode //php bbp_reply_id(); ?> Hi Ivan,
The ‘jumps’ you are seeing are due to different calibration settings being applied to the individual IF paths. Each IF is 20MHz wide and each IF has its own set of corrections. For the TG sweeps we did not optimize the bands for overall flatness, as once you store thru the correction differences are stored in the offset and the trace should be flat. Your measurements will not be affected as long as the store through has been performed.
Regards,
A.J.
AndrewModerator
Andrew November 2, 2015 at 8:50 am in reply to: zerospan and preset question //php bbp_reply_id(); ?> Hi Alex,
Thank you very much for the feedback. I will try to resolve the FM deviation scale issue for the next release. This is a bug that many others have mentioned now. Also thank you for mentioning the preset issue. I will add it to the request list and hopefully be able to address it soon.
Regards,
A.J.
AndrewModerator
Andrew November 2, 2015 at 8:26 am in reply to: Spikes after firmware update on BB60A //php bbp_reply_id(); ?> Hi Donfrank,
Please contact us at 1-800-260-TEST. I am at extension 2. There are a few reasons a firmware update might have failed in this fashion. We can talk about what we can do to resolve this. We are available to talk between 8-3 PST. If you are unable to call during that time, email me at aj@signalhound.com and we can continue this discussion via email.
Regards,
A.J.
AndrewModerator
Andrew October 29, 2015 at 3:46 pm in reply to: Video trigger in modulation analysis //php bbp_reply_id(); ?> Hello Ivan,
Excellent feedback and suggestions. We are aware of the limitations of Zero-Span and hope we can address them in the future. We are also aware of the need for triggering in modulation analysis. As always, I cannot specify any time frame for when we might address these issues, but we do record all suggestions and feedback and use them when making decisions about what to pursue next. We very much appreciate it.
You might be interested to know that there is very basic I/Q capture functionality in Zero-Span mode along the bottom of the screen. It allows you to save a burst I/Q capture to a file as CSV. You would then need to convert it to an amplitude/phase/frequency value in a spreadsheet program, but this might help you.
Regards,
A.J.
AndrewModeratorHi cpitcher,
The VSG operates on USB 2.0. Fortunately USB 3.0 is backwards compatible so you are able to connect the VSG with supplied cable to a USB 2.0 or USB 3.0 port on your PC and everything will work.
If you have any further questions, let me know.
Regards,
A.J.
AndrewModeratorHi Jared,
Thanks for the reminder. This is one that is requested a lot. I’ll try to get this in the next release.
Regards,
A.J.
AndrewModerator
Andrew October 21, 2015 at 9:10 am in reply to: Movement of Legacy API with the tablet //php bbp_reply_id(); ?> Hello zhubajei,
I’m assuming you are targeting the same architecture on both PCs? If you are using Labview 32-bit on one PC and 64-bit on the other, that might be an issue. The error message you have received, “SH_API.dll not found” is a big red flag. This message has to be resolved first. It is possible the Labview configuration has changed on the tablet PC and Labview is looking for the DLL in a different location. (i.e. Different path variable, typos in your path?) My guess is that if you resolve this error, it will work as expected. If both PCs are running a full Windows 10 OS, there should not be any reason one would work and the other wouldn’t.
Regards,
A.J.
AndrewModerator
Andrew October 20, 2015 at 8:59 am in reply to: P25 and C4FM Analysis for Spike //php bbp_reply_id(); ?> Hi Rob,
Thanks for the follow up. With the 4FSK addition to the Spike software you should be able to analyze C4FM signals. My knowledge is cursory, forgive me if I am incorrect, but the C4FM spec is a 4FSK signal with 4.8k symbol rate, 1.8k deviation and root raised cosine filter. Our software is capable of measuring this signal. With our software you should be able to measure modulation fidelity, frequency offset, channel power, and ACPR. What other types of measurements are relevant to you for P25?
Regards,
A.J.
AndrewModeratorHello Witaliy,
Please send me an email at aj@signalhound.com with this message and I can send you the calibration file and instructions.
Regards,
A.J.
AndrewModerator
Andrew September 30, 2015 at 1:03 pm in reply to: Can't Lower RBW/VBW with some spans //php bbp_reply_id(); ?> Hello dwc,
The msvcp110.dll file is part of the Visual Studio 2012 redistributable files which can either be downloaded here http://www.microsoft.com/en-us/download/details.aspx?id=30679 or are part of the Visual Studio 2012 install. If you do not use Visual Studio 2012 you will still need to download these to program to the API. We have 32 and 64 bit versions of the API, so you could download both x86 and x64 versions of the redistributable for your PC. We used Visual Studio 2012 to develop the API which is why you are seeing this.
Regards,
A.J.- AuthorPosts