Justin Crooks November 7, 2018 at 9:33 am in reply to: BB60C temperature Range //php bbp_reply_id(); ?> Serpi,
Yes, we use an environmental chamber to test the units from -40 to +65 ambient for our correction data across temperature. Typically, the internal temperature runs around 8-9 C above ambient in the chamber, so the device corrections should be valid from about -32 to about +74 C internally reported temperature.
Justin CrooksModerator
Justin Crooks October 18, 2018 at 10:02 am in reply to: Low frequency noise measurement //php bbp_reply_id(); ?> I would add two resistors:
1) A 49.9 ohm series resistor for a 50 ohm output impedance
2) A 10k shunt resistor to ground after the DC blocking cap to bleed off residual DC.
Justin CrooksModerator
Justin Crooks October 18, 2018 at 9:55 am in reply to: Compare SA44(A) to SA44B //php bbp_reply_id(); ?> The SA44(A) has no preamplifier, poor spurious performance, and more limited RBW / VBW options depending on span.
The SA44B has gone through several improvements over the years. I would recommend getting something manufactured in 2013 or later for best performance.

Justin CrooksModerator
Justin Crooks October 17, 2018 at 5:03 pm in reply to: Max RF pwr into SA44B //php bbp_reply_id(); ?> You can check Mini Circuits for an acceptable coupler. You might have to get 2 or more to cover your frequency range. Look for connectorized models.

Justin CrooksModerator
Justin Crooks October 16, 2018 at 10:20 am in reply to: Max RF pwr into SA44B //php bbp_reply_id(); ?> Dave,
A few things:
1) The attenuator we sell is only rated for 2 watts. You would need a 10-20 watt attenuator.
2) Typically a directional coupler is used to monitor transmit power. This way, you only lose a small amount of power and you can monitor it while in use. Depending on the “coupling loss (typically 20 dB) you may want additional attenuation. I would recommend aiming for between -10 dBm and 0 dBm to make decent measurements.
Justin CrooksModerator
Justin Crooks October 10, 2018 at 9:55 am in reply to: Generate AWGN with VSG25A //php bbp_reply_id(); ?> There are a couple of options here. You could load an AWGN file as an arbitrary waveform. But the better way for most applications would be to use 1001 tones of random phase, spaced 10 kHz. This will have a CCDF curve that closely mimics AWGN, and will be flat across the desired 10 MHz band as it will automatically apply an inverse sinc correction.

Justin CrooksModerator
Justin Crooks September 10, 2018 at 11:29 am in reply to: External Trigger not working //php bbp_reply_id(); ?> bittware,
Real-time or zero span mode will both capture a 2 us pulse. For zero span mode, use video triggering. For real-time mode, you will want to use the maximum resolution bandwidth (RBW) and even still, the amplitude will read low due to the pulse width being less than the RBW time constant.

Justin CrooksModerator
Justin Crooks August 20, 2018 at 10:32 am in reply to: Compressed IQ stream //php bbp_reply_id(); ?> Preliminary tests on the compressed I/Q streaming indicated that we would end up with fewer channels than we had hoped for, and the interface would be prohibitively complex for most users, so we moved away from it, towards just adding additional memory for full 160 MHz I/Q captures.
Next year, we will have units available with this expanded memory. With video or frequency mask triggering, it should be straightforward to capture and analyze Bluetooth packets.
Justin CrooksModerator
Justin Crooks July 25, 2018 at 10:44 am in reply to: Low frequency noise measurement //php bbp_reply_id(); ?> The other thing to note is your circuit should have a 50 ohm output impedance. If it does not, the SA44B will load it down and the noise reading will be much lower.

Justin CrooksModerator
Justin Crooks July 25, 2018 at 10:06 am in reply to: Low frequency noise measurement //php bbp_reply_id(); ?> Jubayer,
Setup for this type of measurement would be something like:
RBW 100 Hz, VBW 10 Hz, detector settings power, average. Center frequency 7 KHz, span 10 kHz. Reference level -50 dBm.
A 3.3 uF DC block should be reasonably flat down to 3 kHz or so.
Let me know if this works for you.
Justin CrooksModerator
Justin Crooks May 24, 2018 at 9:53 am in reply to: Signal "falling down screen" with low signal levels //php bbp_reply_id(); ?> Ed,
Depending on your serial number, there is a design flaw where under certain conditions the DC blocking cap on the preamp builds up a charge and blocks all signals. Any return path to ground for current will remedy this (including a non-zero attenuator setting).
We can issue an RMA to repair this as well (we add a small resistor). Contact support at signalhound dot com for more information.
Justin CrooksModerator
Justin Crooks May 14, 2018 at 10:57 am in reply to: TG 124A with BB60A, AND suitability to use with other gear for immunity testing. //php bbp_reply_id(); ?> Julian,
The BB60A will need firmware version 5 to be compatible. You can download it here: https://signalhound.com/support/product-downloads/bb60c-bb60a-downloads/The TG124A being used as a standalone generator, and amplified/modulated appropriately, should probably work for immunity testing. Depending on your needs, you may wish to low pass filter the TG output as it tends to have higher harmonics than a dedicated signal generator. You would probably want to automate it using our TG standalone API, which you can get from https://signalhound.com/products/usb-tg124a/

Justin CrooksModerator- This reply was modified 7 years, 7 months ago by
Justin Crooks.
Justin Crooks May 11, 2018 at 10:44 am in reply to: GPS recognition doesn't work //php bbp_reply_id(); ?> AJ,
I remember running into this before. I think this is the bug fix:
https://stackoverflow.com/questions/11775185/open-a-com-port-in-c-with-number-higher-that-9
Also, here:
https://support.microsoft.com/en-us/help/115831/howto-specify-serial-ports-larger-than-com9
Justin CrooksModeratorMaksym,
The power of 2 downsample factor sets the theoretical bandwidth that is available. For example, a downsample factor of 4 gives you 10 MSPS, which theoretically represents 10 MHz of bandwidth. But you should leave 20% for filter rolloff, etc. so generally bandwidth will be 40% to 80% of the total bandwidth based on your downsample rate.
Justin CrooksModerator
Justin Crooks April 20, 2018 at 10:15 am in reply to: SA124B Minimum phase noise //php bbp_reply_id(); ?> Using our legacy API, the device can be configured to use the 2.9 MHz IF (hung mixer mode). You would have to use a higher frequency than 5 MHz (something like 15 MHz at least). I’m not sure it would get you significantly better numbers, but it might give you a few more dB.
There is probably enough RF-to-IF leakage in the mixer that you could use this mode (SHAPI_Configure, use 2.9 MHz IF set to true) and inject your signal directly at 2.92 MHz (at a higher amplitude). This is the closest to direct ADC access the SA series offers.
These are just my thoughts, and may not work, but if you give it a try let me know what you find. The BB60C has more direct ADC access, but realistically, the SM200A would be the ideal choice.

Justin CrooksModerator
Justin Crooks April 16, 2018 at 10:07 am in reply to: BB60C+TG44A strange behavior //php bbp_reply_id(); ?> The BB60C has a built-in DC block. We correct amplitude down to 9 kHz, but there is a lot of VSWR below 100 kHz, making scalar measurements difficult

Justin CrooksModeratorMehran,
We typically need to evaluate the market demand for something like this before we commit to implementing it. And then there is a timeline for feature requests. It rarely happens in a few days unless you have placed a very large order and it is a very simple feature.
If other users chime in requesting a similar feature, the priority gets bumped up. Otherwise, we will evaluate it and make a decision as time permits.

Justin CrooksModerator
Justin Crooks April 2, 2018 at 9:53 am in reply to: Phase noise measurement mode //php bbp_reply_id(); ?> Because of the BB60C architecture, its phase noise, especially at low frequencies (e.g. below 1 GHz) is not very good. So we decided to disable phase noise measurements rather than provide poor results.
The SM200A is recommended for phase noise measurements inside 100 kHz offset. The SA44B can make pretty good phase noise measurements as well, especially outside 100 kHz offset.
Justin CrooksModerator
Justin Crooks March 23, 2018 at 3:08 pm in reply to: CPU Specifications to use multiple SA44B //php bbp_reply_id(); ?> Pritam,
Our API supports a maximum of 8 per computer. Keep in mind that the USB high speed bus is a shared bus, so you probably can’t even have 8 SA44B devices all streaming at the same time, unless you were decimating the data heavily.A possible setup, if you need 24 USB-SA44B’s simultaneously streaming to one PC, would be to have 7 PCs in a network. Use 6 PCs to connect to 4 SA44B’s each, and each stream their data to the 7th PC. It would be a bit of a project, but any project that requires 24 spectrum analyzers is going to be a bit of work anyway.

Justin CrooksModerator
Justin Crooks March 19, 2018 at 9:48 am in reply to: Amplitude modulation with VSG25A //php bbp_reply_id(); ?> Yes, the VSG25A would work well for this.
- This reply was modified 7 years, 7 months ago by
- AuthorPosts