Forums › BB Series Discussions › BB60C temperature Range
- This topic has 4 replies, 4 voices, and was last updated 6 years, 2 months ago by Justin Crooks.
- AuthorPosts
SerpiParticipantHello,
The BB60C with extended temperature is given for -40 to 65’C, but what is the link between the external temperature and the read one on Spike ?
What is the offset ?
AndrewModeratorHello Serpi,
The temperature read in Spike is the internal temperature of the device. The BB60C is not able to provide external temperature readings. When we calibrate the BB60C across temperature we use environmental chambers to control the external temperature.
Regards,
Andrew
SerpiParticipantOk, I search a way to present an alarm to protect the BB60C in case of high temperature, what is the maximum for internal temperature in this case.
I suppose that you have tested in your environmental chambers at 65’C and -40’C and have the internal temperature results ?
TishersParticipantWhen you let the unit run for hours at a time the entire case gets quite warm. As with any test equipment you would want to give the analyzer some time to warm up to minimize drift.
the -40 to +65C specifications are very common for the electronic components.
Justin CrooksModeratorSerpi,
Yes, we use an environmental chamber to test the units from -40 to +65 ambient for our correction data across temperature. Typically, the internal temperature runs around 8-9 C above ambient in the chamber, so the device corrections should be valid from about -32 to about +74 C internally reported temperature.- AuthorPosts