kaiserParticipantkaiser March 31, 2026 at 7:20 am in reply to: SM200C Streaming IQ – no Triggers //php bbp_reply_id(); ?>
Thanks! It’s readily apparent that my body was no longer on vacation, but my mind still was. I appreciate the quick response here.
kaiserParticipantI wanted to come back to this — sorry, I had some work travel and haven’t been able to finish off updating y’all and doing some testing.
What we found:
The SM200B’s *did not* have this synchronization issue. When we kick them both off at the same time, with the same 10MHz reference and queue scans, it can run for hours and still be 1:1 in sync with scans.
With the SM200C’s, we notice a divergence right away. I assume with the latency inherent in networking, we end up with multiple scans on one done before the other even gets the command. Makes some sense. But I would’ve thought that if we queued the same # of sweeps and finished them that we’d see identical numbers at the end, which we aren’t. It might be an issue with our trigger line for counting sweeps?
So we went to try and work around it, but just doing the single sweeps at a time. Which works for us, and seems to count correctly (potentially exonerating our trigger line? If it counts correctly here, then it should be able to count queued sweeps too, right? The queued sweeps are ~20% faster though…but that’s fine for us for now).
I think that we might need to scope the GPIO outputs during these sweeps to figure out what’s going on.
kaiserParticipantWe have the two SM200’s together on the same 10MHz reference (one outputs it, and the other inputs it), and it seems to keep them well sync’d.
Two separate triggers go into the FGPA from the SignalHounds, and cause two separate things to happen. Now, while the FPGA is off doing one thing for the first trigger, theoretically the second trigger could get missed. But we’ve timed the actions on the O-scope, and the FPGA goes back to listening in under half the time that the signalhound keeps the GPIO high while resetting. So we don’t think the issue is there (although this is one area for us to revisit if the rest don’t turn anything up). But again, if we were missing a trigger, we’d expect to never be able to recover from that miss — that we’d always be off by one…so it’s odd.
Right now we cover ~200MHz of spectrum, so IQ streaming isn’t feasible, especially since we do have a lot of SM200B’s in our legacy and other designs…but we’ve definitely been toying and playing around with it some. It’s nice to have the sweep handed to us with all the FFTs and windowing and everything done, instead of having to do that ourselves. But one day we might bite the bullet and get there (and probably will need to soon for another project we’re using the SM2000’s on).
kaiserParticipantThanks Andrew.
I do have to say that with SM200C’s, the difference in speed between Queued Sweeps and non-queued is basically imperceptible in our application. It’s very very response non-queued…I’ll be able to test how the SM200B’s hold up doing the same. If I remember right the non-queued was significantly slower.
Here’s some of the various code bits for how we’re configuring the SignalHound.
sm_api.smSetSweepSpeed(handle, SmSweepSpeed.smSweepSpeedFast);
CurrentStatus = sm_api.smSetSweepCoupling(handle, 125.0e3, 125.0e3, 0.0001); //Gives a 3.1kHz scan, 138us blanking (similar spectrum sweep time, about 50% duty cycle).
sm_api.smSetSweepDetector(handle, SmDetector.smDetectorAverage, SmVideoUnits.smVideoLog);sm_api.smSetSweepWindow(handle, SmWindowType.smWindowNutall);
sm_api.smSetSweepSpurReject(handle, SmBool.smFalse);
sm_api.smSetPreselector(handle, SmBool.smFalse);//GPIO configuration
int gpiosteps = 2;
sm_api.SmGPIOStep[] Steps = new sm_api.SmGPIOStep[gpiosteps];Steps[0].freq = 2.2e9;
Steps[1].freq = 2.38e9;
Steps[0].mask = 0x00;
Steps[1].mask = 0x0F;sm_api.smSetGPIOState(handle, SmGPIOState.smGPIOStateOutput, SmGPIOState.smGPIOStateOutput);
kaiserParticipant- This reply was modified 2 years, 7 months ago by
kaiser.
Andrew, thanks for the quick and thorough reply.
I’m going to keep investigating this some more myself to try and understand what’s going on and verify the rest of the system. It just *feels* like a Queued sweep bug since any time I go above 1 (basically queueing at all) it starts to happen. But feelings aren’t fact, so we’ll do what we can to finish narrowing down the issue.
– How are you switching between your sources? Are you using the smSetSweepGPIO function?
— No. We have the SM200’s set to produce a waveform (go low when scanning, transition high when near the end of the sweep) and that transition to high then tells the FPGA to move to the next position.
– If yes to the question above, have you done any monitoring of those GPIO pins?
— We monitor counters in the FPGA for that rising edge, and they always seem to line up…but then again, when you’re doing something at multiple kHz, and there’s only ever an off-by-one bug it would be a needle in a haystack.
We don’t think that we are missing triggers because otherwise it wouldn’t go back! We know if we have a bad trigger line because then what was in slot 10 goes to 9 and then to 8 and then to 7, and so on. But this corrects itself! Which is just odd.
– Any way to verify they are cycling through all your states and not missing one periodically? Whether it’s hardware/software, that would be extremely helpful to know.
— Yea, that is one of our concerns. I don’t know why it would go away when we set the queue to one, unless we were for example missing a trigger and then catching up somehow? But I don’t know how the FPGA would do that since we’re just looking for an edge transition on a pin. One of the avenues we are investigating on our end. But again, it only happens when we queue sweeps…one thought would be that the waveform / time the GPIO is high before the next sweep happens is longer when we’re not queuing, so we “catch” it more regularly…but then again, why would it correct itself!?! It would process moving bins in that instance, I would think.
– Am I to understand that the “sweep offset” sticks around for many cycles? So if you are cycling through all 20 slots, are all 20 slots now off by 1 for some duration until it corrects itself? Or is it just slot 10 that shows up in slot 9 for several seconds until slot 9 reverts?
— Yes, we really have a couple hundred “slots” that we cycle through. And we kind of see the whole thing shift one left, and hold there for a bit and then shift one right, back to the original and desired location. Then it’ll go back and forth doing that. Really odd, honestly.
– Does using just 1 receiver make the problem go away? Even with 16 queued sweeps?
— Our hardware and software architecture is very very intertwined with having two SM200’s working in concert with each other. I think we might be able to hack some places in the code to reduce it down to one signal hound, but only for a little bit before buffers and other things start overflowing and getting too far out of sync elsewhere. Otherwise it’d be a huge effort to get it down to just one.
– Is it always a specific slot, e.g. 9/10? Or random?
— It appears to be all the slots kind of shifting by one. Almost like we started getting scans out of order for a bit or something. It seems more like the problem would be elsewhere. But it is 100% solved when we queue only 1 scan, and is apparent when we go for more than one scan. I honestly haven’t been able to come up with a theory as to why / how the queued/non-queued is at play here. But as far as we can tell, it is.
– Does the “sweep offset” happen on both receivers? Or just one?
— It does happen to both. We are looking into that more this rest of this week. We had an issue preventing us from looking at both and seeing if they do it at the same time, or if it’s a random correlation between the two.
– Let’s say you’re switching through slots 8,9,10,11. Does it now look like 8,10,10,11 for 15 seconds and then go back to 8,9,10,11?
It looks like 1,2,3,4,5,6,7,8,9 goes to 2,3,4,5,6,7,8,? for a while and then back to 1->9 in order. We’re investigating whether the ? is just old data in there, or if it’s 1 from the previous slot. It’s actually fairly hard for us to put energy *only* in one slot with our current setup, but figuring this out I think might provide a lot of insight, and that’s what we are attempting today.
It’s a tough nut to crack, because the entire system is orchestrating this at multiple kHz, and there’s a lot at play here, and we’ve tried a ton of things across all the other components, both hardware and software at play here, and it wasn’t until we only queued 1 sweep up did anything fix it.
With queued sweeps, is there maybe some extra allocation / deallocation (and thus garbage collection or other process) going on under the covers in the driver that’s maybe going on here?
kaiserParticipantJust for reference it worked flawlessly when we used the I/Q recorder utility.
kaiserParticipantkaiser September 13, 2023 at 11:50 am in reply to: VSG60 Streaming Playback //php bbp_reply_id(); ?>
Thanks. Super helpful as always Andrew.
kaiserParticipantkaiser September 13, 2023 at 10:42 am in reply to: VSG60 Streaming Playback //php bbp_reply_id(); ?>
Thanks. We’ll investigate the network and SSD performance a bit. I don’t remember seeing an issue shown on the display, and we were looking for one.
The SM200C and VSG are on different computers, and the recordings are 20 seconds long. We are using the Streaming mode on the VSG60, and we aren’t doing much else with the computer, and that’s also why we tried bumping it down to 20Ms/s and 5MS/s playbacks, and we kept seeing the data error bursts (but on a correspondingly longer timeframe before the bit errors; so like at the same point in the file, which is why we’re goin to investigate the recording a bit more).
We were setting the SM200C into zero span mode, and then using the I/Q recorder thing on the bottom since that also produces an xml file with the recording details. Do the error labels show there? We can also try from the I/Q recorder dialog thing and see if its different; we just don’t get the XML file with the details. Is that by design?
kaiserParticipantThanks Justin. Great and prompt answers, like always.
That latency looks like it’ll probably be in the noise to me, so we’ll go ahead and plan on trying out the C model.
kaiserParticipantkaiser November 1, 2021 at 10:40 am in reply to: SM200B vs SM200C Questions //php bbp_reply_id(); ?>
Thanks Justin, I appreciate the info. Glad to hear that it shouldn’t generally be an issue (we usually have a bit of extra processing oomph, and we’ll just make sure we’re good moving forward too).
Can you share typical expected latencies (and what the difference is from the B to C)? I’m very latency sensitive, so am wanting to test some new stuff on the C and see if I can tighten some loops if it’s usually a bit better.
kaiserParticipant- This reply was modified 6 years, 2 months ago by
kaiser.
Thanks Justin. I had tried all of those things before posting with no joy. I tried the 12.0 version, the .28 and the .26 version of their drivers, uninstalling, re-installing, etc with no success. Using the CDM uninstaller and more.
It’s a RS-422 chip for the VN-310, and I wonder if there’s something special there or not.
I *swear* that I had all 3 working together about 4-5 weeks ago, and I’m not sure what broke it — a Windows update, or something else, or if I just fooling myself about all 3 working at once. We don’t even need to pull up Spike, just plugging a single SM200B into power, is an immediate stoppage on the VN-310….like right as Windows enumerates the drivers, it goes bye-bye. I’ll be watching data stream in the terminal, and then it will stop and just send NULL and BREAK repeatedly. Unplug the SM200B and instantaneously data starts flowing again.
We bought some converters so that we can try and shove it in a resident serial port on the computer and see if it works that way.
For my own reference/googling/spec sheet hunting, what does the SM200B use for it’s USB chip — is there some serial<->USB conversion or something else going on on the front end?
Thanks!
kaiserParticipantThanks Andrew. I had thought we might have to do a handoff, and that’s like plan D. I’m still working through A-C, and seeing if there are any sub-plans to implement.
The times you gave for reconfiguration and super helpful and will help me put together some trade-space regarding how agile (I’ll toss in some margin) we’d be able to do different schemes.
kaiserParticipant- This reply was modified 6 years, 7 months ago by
kaiser.
Looping pins 3 and 11 together did bump the output signal to 3.3V.
It was crazy — I have a system that’s supposed to trigger at 1.4V, and I could see the ~1.6V SH level going into it, and it just wasn’t triggering. If I slowed it down, it would sometimes trigger. Very odd. I hooked up a function generator and exactly matched the SH output signal, and it triggered perfectly!
So, something about the pull down and series resistor or amount of current available to source coupled with my digital IO card was causing more voltage loss inside the card. The bump to 3.3V in solved it all.
kaiserParticipantThanks Justin, we’ve had this going and working for a while. We’ve been playing with sensitivity some, and we’re only triggered our device by the 15mV 🙂
If I wanted to up the drive voltage out, do I supply a new Vdd in? Is there some circuitry already setting that somewhere? It would seem so since there’s obviously some drive voltage out already 🙂
kaiserParticipantkaiser January 16, 2020 at 7:44 am in reply to: Another Synchronization Question //php bbp_reply_id(); ?>
Thanks Justin. That was my worry 🙂
It makes things quite a bit more complicated code-wise on my end, but I think still manageable…It would be nice if I could queue and get everything ready, then trigger both to start the queued sweeps electrically, rather than via software. I could stop every now and then for a re-alignment after drift due to them just going full bore.
My bandwidth I’m monitoring is ~200MHz, so unfortunately streaming options are out for me…
kaiserParticipantUpon further investigation, you can change the sm_api.cs to just be like:
[DllImport("sm_api", CallingConvention = CallingConvention.Cdecl)] public static extern SmStatus smGetDeviceList(int[] serials, ref int deviceCount); [DllImport("sm_api", CallingConvention = CallingConvention.Cdecl)]Then if you’re on Linux, it looks for sm_api.so, libsm_api.so and on windows it looks for sm_api.dll
Nice little benefit for interoperability. If you need to specify x64, x64 and other arch’s, you’ll probably need to SetDllDirectory or LoadLibrary with a dynamically added name, or a custom build step to rename the proper dll and put it with the assembly. DllImport only can do so much matching.
kaiserParticipantkaiser August 20, 2019 at 8:04 am in reply to: What is wrong with my measurments //php bbp_reply_id(); ?>
So, have you tried a free-field measurement yet (aka not in the chamber)?
Antenna test is intricate. It’s hard to truly say what is or is not the problem here because there can be dozens of things that cause something like this, or it could really just be that.
Just looking at it, given the frequencies you’re talking about, if an engineer came to me asking about this and what to do next. I’d suggest a free-field measurement. In general, my comments would be that you’re testing at a low frequency (250MHz) has a ~4′ wavelength. In order for the fields to enter into the antenna, you’re going to want other things (walls, absorber, etc) at least that far away from it unless you’ve specifically verified the test setup, that you have absorber behind the antenna to prevent reflections, but far enough away to not artificially couple and affect the antenna. There are a lot of things much closer than that at the frequencies your talking about. Also, the chamber could also end up acting more like a resonant cavity than a chamber depending upon size and exact geometry. Again, maybe not a problem, but could be. Also, have you ran S11 on the antennas post install? Are they being detuned, or a reflection coefficient issue created during install? And so on.
The bottom line is that from the pictures, given the issue that you’re seeing, you could spend days/weeks/months tinkering with the test setup and never hit a definitive conclusion on the issue, as there are potentially many, and they could interact in complex ways. So, I’d tell my engineer to check S11 an as-installed items, and check it in various chamber configurations. S11 open-air vs chamber should be the same; if substantially different, something is up.
Then separate out the test chamber from the measurement, and then add pieces back in. Go find some open space somewhere, and just run a real quick simple test. Collect that data, and then start introducing your test setup piece by piece, verifying that you’re good every step as you do so. You’ve already half did that with the test from a certified house, but it’s just hard to truly say you’re replicating or there’s an antenna problem, etc when all we really have with these exact items is a somewhat problematic test chamber setup.
kaiserParticipantkaiser August 19, 2019 at 9:03 pm in reply to: Queued Sweep Synchronizations //php bbp_reply_id(); ?>
[StructLayout(LayoutKind.Sequential)]
public struct SmGPIOStep
{public double freq;
public Byte mask; // gpio bitspublic SmGPIOStep(double f, Byte m)
{
this.freq = f;
this.mask = m;
}
}[DllImport(“sm_api.dll”, CallingConvention = CallingConvention.Cdecl)]
public static extern SmStatus smSetGPIOSweep(int device,[In, Out]SmGPIOStep[] steps, int stepCount);I just downloaded the latest API, and was running it against some of my code. I had to edit the sm_api.cs again to add this in, so that I could set up GPIO Sweeps from C#. It seems to work great again. This is the 3rd project where I’ve had to make this edit, as suggested here. Is there any reason why this bugfix hasn’t been included in the sm_api.cs file yet?
kaiserParticipant- This reply was modified 7 years, 5 months ago by
kaiser.
kaiser April 26, 2019 at 8:50 pm in reply to: BB60C, EMC measurements:How to lower the noise floor of my setup ? //php bbp_reply_id(); ?>
Yea, I understand where you’re at regarding the noise floor. The thing is that you’re just continually bouncing waves in the chamber, which will build up (you’re in a microwave oven) and increase the noise floor.
Think of the difference in background noise in a small library or church that’s all tile, granite and glass. You can hear everyone doing everything, and there’s a small “din” as all the sound bounces around with nowhere to go. You’re noise floor is higher. Take that same building with the same people, hang some tapestries, put that carpet down and it will be tens of dB quieter and without that “din”. You don’t need to upgrade the whole chamber, but some basic foam like I said can help a lot.
For the LNA, you don’t need anything fancy. I don’t know your frequency range and necessary specs, but we typically would use something like this, maybe even two of them depending upon line loss and the gain budget:
Put it as close to the receiving antenna as you can (right on the connector if possible), so it drives the signal down the cable and into the BB60. This should increase your SNR.
If you’re already near max input power level, the LNA could cause some issues there. You may want to toss a filter in in front of the LNA to knock down whatever else is causing the issue. But some red warning light in my head is going off because what’s being described isn’t totally matching with what I’ve seen in the field. The BB60C should have more than enough dynamic range for nearly any test. If you’re driving it near IF overload, but you’re still having sensitivity problems as evidenced by the attenuator test, then it feels like something else is going on. Have you checked for other interfering signals, zooming way out on the Spec An? Maybe your chamber isn’t as tight as you thought, or there’s some injection leakage. Is your test equipment *outside* the chamber, and just the antennas inside? Is the chamber appropriately sized for your lower frequencies (too small and it will act more like a waveguide)? Or maybe the chamber is just bouncing the signals around too much and you’re getting to much of that echo bouncing around nailing you. Have you tried it outside the chamber with the power tuned down some?
A box of anechoic foam (4 pieces) isn’t too expensive, and clever placing of it can help a lot. We don’t always use a chamber (in fact we often don’t), like I said slap a couple of pieces on plywood and make something that looks like an open box laid sideways that the antenna fits in and you’ll knock out the vast majority of interference. Add in a height differential or a piece of foam at the halfway point on the ground, and you’re pretty good to go.
So, try the LNA, that will definitely give you more sensitivity for better measurements. But my guess is that there is another gremlin or two out there that you’ll need to slay to get good results. It probably will be something you would never have thought of. Keep at it, RF test is hard!
edit: You also could always have a bad device; but that’s one of the gremlins to slay that I was talking about. Get some of the other stuff figured out, and see where it leads you.
Bypassing open air and just connecting Tx (or an appropriate surrogate) to Rx with an appropriately large attenuator (start with lots for safety, then dial it back) should give you the best signal characteristics and can help determine if you have a bad device or not. Could also be a bad Tx that’s putting out broadband noise also… Then you can also add in components walking the chain to see where things change. And antenna / atmosphere is a component in your link budget, so eventually you add that back in also 🙂
kaiserParticipant- This reply was modified 7 years, 5 months ago by
kaiser.
Thanks Andrew. I’ll play with it once we order the SM200’s since it’s not currently a pressing issue for us. Just exploring potential risks.
Our processing is also in other threads, and we’re seeing pretty low utilization compared to what we expected so we might be ok. I can write a simple program to check that it’s working or not, but it’s always useful to make sure that you’ve set the right permissions on the actual final executable process 🙂
I updated our current script to use #1 above, set the process to -10 nice and was able to execute it without having to sudo. htop showed it at -10 nice level, and it did seem to get first priority over CPU resources. Used -20 (realtime, aka all the CPU) and it did consume all of the CPU. So it should be easy to test once I get to that point.
But reading more about it, niceness probably isn’t going to work.
The easiest way would be to allow your application to run as root by editing the sudoers file to allow it without password. But then you’re giving all root capabilities to the .exe, which if altered could result in complete pwnage of the system.
Next easiest woul be to allow access to a command, like set_rlimit
or chrt or setcap() via the sudoers file. That should be simple and safe, but I’m not in a place to test it right now. Linux just has so many darn ways to skin a cat, it can be hard to figure out which one is the best.- This reply was modified 2 years, 7 months ago by
- AuthorPosts