ANguyenParticipantHello!
I am currently an undergrad student and have just started working with the SA124B and TG124A. I am trying to create a virtual instrument using LabVIEW. However, I keep running into this call chain error, even when running the example VIs provided by SignalHound. I have attached a screenshot here of the error that I am encountering with the SA Series Sweep Continuously VI.
I have tried to replace the drivers as well, following the official tutorial found online. I have attached pictures of my driver folders.
Sorry for the lack of detail. Let me know if I can provide anything else that would be of help. Any help would be much appreciated. Thank you!
Attachments:
You must be logged in to view attached files.
RogerModeratorHi ANguyen,
Thanks for the details provided.
The error is -8: Device Not Found. The error codes can be looked up in the API docs.
Is the SA44B working in Spike? If so, make sure Spike is not running when attempting to connect in LabVIEW. If not, it would be best to troubleshoot that first.
Note that the USB driver needs to be installed separately. It can be downloaded with this link. Unplug devices, right click on the driver and select “Run as administrator”. Then plug in device and run Spike.
If you are not able to get the SA44B running in Spike, please reach out to support@signalhound.com for more detailed troubleshooting.
– Roger