osyParticipantDear Justin,
Thanks for your response.
I’m attaching screen capture from TG124A user manual.
1) It says -30~-12dBm (typical -30dBm~-10dBm)
2) Manual says 100KHz~12.4GHz. So I translate it to -10dBm over -30dBm~10dBm. So I tested it @-10dBm as max.1. Please let me have your kindness once again what am I wrong?
2. One more thing, the calculation from thermal sensor is very tricky as you said. do you have any reference material so that our team can discuss about that?
(Because we’re failing to get certi in Korea, we’re serious to check amplitude spec items…)Thanks,
SYAttachments:
You must be logged in to view attached files.
osyParticipantosy January 28, 2021 at 3:30 pm in reply to: Discontinuity issue in Scalar Network Analysis using SA44B + TG44A //php bbp_reply_id(); ?>
Dear Justin,
Thansk for your comment.
And can you give me any valuable option try in API or configurtion/setting to smoothnout? Even 6dB attenuator shows similar result even though it align a little bit.
(image rejection, reference level,…)
End customer push me to investiagte any resolutuion withoout sacrificing dymanic range.
Thanks,
SY- AuthorPosts