AndrewModeratorHello Gerhard,
1) We have two graphical ways to control the TG, either through the Spike software and through the application TGStandalone.exe found in the Spike installation directory at C:/Program Files/SignalHound/Spike/. Both ways only allow you to set the amplitude and frequency manually.
2) The output level on the TG is between -10 and -30.
Outside of the scalar mode, the TG is a simple CW generator. Yes, the scalar mode is primarily built around return loss and insertion loss measurements, filters and antennas being two of the larger use cases.
If you want to talk in more depth about your measurement requirements offline, you can email us at support at signalhound dot com.
Regards,
Andrew
AndrewModeratorAndrew July 21, 2017 at 1:31 pm in reply to: USB Serial Drivers do not appear to load //php bbp_reply_id(); ?>
Ron,
Please call us at 1-800-260-TEST Ext.3 or email us at justin at signalhound dot com.
Regards,
Andrew
AndrewModeratorAndrew July 21, 2017 at 11:57 am in reply to: USB Serial Drivers do not appear to load //php bbp_reply_id(); ?>
Hi Ron,
Let’s try from the beginning. There is no need to reinstall Spike. Just grab the latest USB driver from this link.
https://signalhound.com/sigdownloads/Spike/CDM%20v2.12.00%20WHQL%20Certified.exe
When you install it, you will want to run it “As Administrator”, you can do that by right-clicking the installer and selecting “Run as Administrator”.
Ensure the Signal Hound device is not connected when you install the driver. The first time you plug it in after successfully installing the driver, it might take a few seconds for Windows to recognize the device.
If you get the solid green LED but Spike still doesn’t recognize the device, look at the FAQ on this page http://www.signalhound.com/Spike at the question “The Spike software reports “Device Not Found” (SA44/SA124/TG44/TG124 only)”.
I look forward to your response.
Regards,
Andrew
AndrewModeratorAndrew July 20, 2017 at 8:27 am in reply to: Cannot pause spectrogram on Spike controlling BB60C //php bbp_reply_id(); ?>
Hi Gary,
You can stop the spectrogram by single stepping the measurements, by pressing the “Single” button in the toolbar. This will stop both the sweep and spectrogram. There is no way currently to stop just the spectrogram.
Also, keep an eye out in the next 2 weeks for version 3.1.8 where we will have revisited the spectrogram display, giving you the ability to scroll through past history (up to several thousand traces) and view the amplitude/frequency/time of any event in the spectrogram.
Let me know if you have additional questions.
Regards,
Andrew
AndrewModeratorHello Abaker,
The USB drivers we link in the FAQ are specifically for the various Intel CPUs that support USB 3.0. If you are using an external card, then those drivers won’t be applicable. We recommend a class of Intel CPUs (4th generation or later i5/i7’s) because they implement USB 3.0 support in hardware that has proven to work at the sustained throughput we need (Typically referred to as native USB 3.0 support). We have a poor track record with external cards and third party USB 3.0 vendors such as Renesas and ASMedia.
Since the card has worked for you in the past, it might be that you need to find the latest drivers for the card. Or it might be the card isn’t going to work in combination with this PC. The Xeon processor you linked was made before the Native USB 3.0 push from Intel and widespread USB 3.0 adoption on PC hardware.
Regardless, both the CPU and the external card are outside the minimum recommended PC requirements, and I suspect this is why you are seeing the “Device Not Found” error message.
Let me know if you have additional questions.
Regards,
Andrew
AndrewModeratorAndrew July 9, 2017 at 11:50 am in reply to: Cannot connect SA44B with Spike //php bbp_reply_id(); ?>
Hello Causilgey,
Just to clarify, you have to go into the properties menu to find the advanced tab after right clicking on the Converter A/B items?
Please contact us via email for additional troubleshooting. Please email me at aj at signalhound dot com. Additionally please CC justin at signalhound dot com. We will help you determine the source of your issue.
The X240 is capable of running the SA44 and BB60.
Regards,
Andrew
AndrewModeratorHi gga33,
Thanks for contacting us. I can help you with your questions.
a. In our software switching between sweeps and QP detector is manually done with the Start/Stop buttons in the control panel under the “Meter Settings” tab. Additionally in the Peak Table control panel along the bottom of the screen, you can select a peak and move the settings to the QP detector by pressing “Selected Spur to Meter”. In the meter settings you can control the dwell time parameter.
b. The only way to change the measurement points in the sweep is to change the RBW. You currently cannot directly set the number of points in the sweep.
c. You will need to use the “Scalar Network Analysis” measurement mode in Spike for return loss measurements. There is a step-by-step setup guide in the manual (you can bring the manual up in the help menu). You can also watch the Justin setup up the software for SNA measurements in this video https://www.youtube.com/watch?v=hxUZx-9DMYg
Let me know if you have follow up questions.
Regards,
Andrew
AndrewModeratorAndrew July 6, 2017 at 10:01 am in reply to: Cannot connect SA44B with Spike //php bbp_reply_id(); ?>
Hi Causilgey,
I just verified that on Windows 10 Pro the Load VCP checkbox is still present for the Serial Converter A/B USB entries. If you do not see them or the advanced tab, I would wonder if you have installed the USB drivers on the Win10 machine? You can find the installer on the Spike download page or in the installation directory on your PC.
Take a look at the FAQ on the Spike download page, https://signalhound.com/spike/ See the question titled, “The Spike software reports “Device Not Found” (SA44/SA124/TG44/TG124 only)”. Follow the troubleshooting steps there.
If you are still unable to get the device working, contact me directly at aj at signalhound dot com.
Regards,
Andrew
AndrewModeratorAndrew July 5, 2017 at 9:49 am in reply to: SA44B (ARM / Pi2 version) for spur hunting //php bbp_reply_id(); ?>
Hi Ron,
I would consider a small windows PC or laptop for best use. There are several restrictions on the RPi2 platform due to driver and throughput issues we have over USB 2. The PC restrictions are very low for the SA44B if you are going to use the API (most PCs will have no problem)
Regards,
Andrew
AndrewModeratorAndrew July 5, 2017 at 9:43 am in reply to: Understanding SA44B Limitations – Windows/Linux/ARM //php bbp_reply_id(); ?>
We have not tried the latest FTDI ARM drivers, nor have we tried the XU4 platform.
Regards
AndrewModeratorAndrew June 30, 2017 at 9:18 am in reply to: BB60C GPS Timestamp Receive Problem //php bbp_reply_id(); ?>
Hello blackmore90,
What happens if you call the bbGetIQ function a few more times, (call it in a loop for a few seconds) I believe it requires at least 1 PPS signal from the GPS before it can start accurately time stamping data.
(And I’m assuming you have checked the return values of the functions to verify they haven’t thrown any errors)
Regards,
Andrew
AndrewModeratorHello Raash,
Thank you for the feedback. At the moment, we will continue to ensure the old software version is available for access to the smith chart. The older software can be installed side by side with Spike. We do not have immediate plans to include this in Spike. I apologize for the inconvenience.
Regards,
Andrew
AndrewModeratorSounds good David.
Additionally, in case you weren’t aware, we have C DLL programming interface for our devices, if you wanted to retrieve IQ data directly from the receiver programmatically, you can do that. The API files are found on the devices respective downloads page, you will find an IQ streaming example in the downloads folder.
Regards,
Andrew
AndrewModeratorHi Charles,
If simply re-saving the file fixed the issue, I’m going to guess that the line endings is what is resolved it. Our software probably saves newlines with ‘\n’ and excel will replace them with ‘\r\n’. After looking at the files, the resaved version also removed the spaces between the values on each line, maybe that?
Glad you found a solution,
Regards,
Andrew
AndrewModeratorHello mpatino,
I hope I understand your question correctly.
The SA44B requires a solid USB connection. Any latencies or time outs will cause the software to stop working. This is because the software is requesting large amounts of data from the receiver and any gap or loss in the data will result in invalid measurements. If there is too much latency the device will also likely not even be able to initialize.
Regards,
Andrew
AndrewModeratorHi David,
We only have a binary format IQ file. You will probably need to write a small bit of code to convert this to human readable format. There is a a section in the user manual about the format, but it is simple enough to describe the basics here. The binary file contains alternating IQ values stored as signed 16-bit shorts that have been normalized with a scalar. The scalar is stored in the companion XML file. You will need to unpack the IQ values, convert them floating point values, and apply the scalar (multiply each IQ value by the scalar).
We have talked about writing a conversion utility to convert our files to CSV, but don’t have immediate plans for this.
There might be other tools like Matlab which could greatly simply this process.
Let me know if you have additional questions.
Regards,
Andrew
AndrewModeratorHi Charles,
Sorry for the late response. I was hoping another IAS user would chime in with their experience. Does the IAS software have a user manual or anything else that would describe the file format it expects? If you can find this, I could probably help determine what would need to be done to make it work.
The file that you said works has several configuration settings included, is it possible the IAS software expects one/some of these?
Regards,
Andrew
AndrewModeratorHi mpatino,
I’m afraid I have never used the Digi USB anywhere products, although I do remember one customer in the past successfully running an SA device on them. Can you verify whether the device is showing up as a USB device in the device manager? Also, is the device properly powered on the Digi hub, i.e. LED is green on the device? I wish I could be of more help.
Regards,
Andrew
AndrewModeratorAndrew June 7, 2017 at 10:34 am in reply to: Calibration software for SA and TG //php bbp_reply_id(); ?>
Hello Ganesh,
Yes, we can perform a calibration of the VSG25A. The cost of calibration is 150 USD (I think). You can call us at 1-800-260-TEST to set up a calibration, or send an email to lori@signalhound.com.
Let me know if you have additional questions.
Regards
AndrewModeratorAndrew June 5, 2017 at 10:30 am in reply to: USB-TG44A not working in Windows, works in Linux //php bbp_reply_id(); ?>
Good news David. Thanks for the feedback as well. We will update the FAQ as soon as we can.
Regards,