yo3fffParticipantHello,
Just discovered an error reading the frequency with Spike from ver. 3.1.4 up to now (3.1.8) using SA44B. Frequency is drifted up by variable amount +1.78kHz for a 10MHz signal and 26.7kHz for a 150MHz signal.
Using the same SA44B and the same external signal but an earlier Spike version (3.0.22), everything is all right. 10MHz is 10MHz, 150MHz is 150MHz.
From the above facts I concluded there is a bug on 3.1.x Spike versions.Look forward to hear about a patch for the bug.
BR,
Cristian Negru

AndrewModeratorHello Cristian,
How are you measuring the frequency offset? Are you in sweep measurement mode and using markers? If so, what is your RBW?
Can you verify the frequency offset using tools such as the Utilities->Frequency Difference meter, or the carrier offset readout in Analog Demod measurement mode? In both measurements you will want to set the center frequency and reference level for your input CW.
I look forward to your response.
Regards,
Andrew
yo3fffParticipantHi Andrew,
I’ve sent the same message to your support e-mail address together with Spike preint screens. I will attache two of theme here too.
Let me explain again the used setup measurement.
An RF generator was connected to SA44B input throught a 50ohm 6dB fixed attenuator. This hardware setup was not changed throughout the test! The RF generator was set to transmit a fixed unmodulated CW frequency. That frequency was 10MHz and 150MHz. I chose two frequencies to show that the frequency offset is not constant but increase with frequency!
Now, what was changend in the measurements was just the Spike software version! The hardware connection was keept unchanged. The RF generator is miniVNA PRO and have very good frequency stability (less than 10Hz/hour).
As can be seen from the attachement, with Spike ver. 3.0.19 the RF signal has a sharp 10.000000MHz and with the latest Spike ver. 3.1.18, the same RF carrier was read as 10.001773MHz. Somehow the new version add about 1.773kHz to the real frequency and this is even worst as the frequency increase.Also from the att. you can see that RBW=VBW and are the same for both versions (30Hz). The analisys mode was Sweep measurement without spurious rejection. Ans yes, I put the marker 1 just to be clear the frequency reading. I use the peak search function which is working well.
I think you missunderstood the problem! There is NO hardware frequency offset. So how can I verify the frequency offset with different software version?Please let me know if now is clear.
If you need more information, please let me know.Thank you for support.
BR,
Cristian Negru
Attachments:
You must be logged in to view attached files.
AndyParticipantHello Cristian
I don’t want to interrupt, but I have the SA44b running on 3.1.8 software and I see no such error. The centre frequencies and the associated errors would seem to indicate a reference frequency error.
Attached is my screen grab. Signal source is HP 8648b.
Good luck.Attachments:
You must be logged in to view attached files.
yo3fffParticipantHi Andy,
Thanks for the advice. Problem solved.
I didn’t know that on the new version 3.1.x the ref. freq. should be adjusted.
For this, as Andrew mentioned, I’d used Utilities>Timebase Adjustment.Thank you both of you!
BR,
Cristian Negru