Forums › BB Series Discussions › Invalid cal data, potentially corrupted
- This topic has 1 reply, 2 voices, and was last updated 22 hours, 16 minutes ago by
Andrew.
- AuthorPosts
zbroyarParticipantHi,
Proud BB60D owner here 🙂
I know it’s just a warning and your docs say “Errors stop the flow of execution, warnings do not”, but I have two questions:
1. Can I continue using the device?
2. How do I get it back to a calibrated state?
AndrewModeratorThis warning indicates that either the adjustment data is corrupted on the internal flash, or there is USB data loss when transferring that data to the PC (which occurs during the open process). You can keep using the instrument, but either scenario would result in potentially incorrect measurements.
To start troubleshooting, let me have you send us some information to support@signalhound.com,
– Is this problem new? or has it always been occuring?
– What is the make/model/OS of the PC?
– Are you using a USB hub?
– Do you have a different USB cable you can try?
– What is the S/N of the instrument?We look forward to your response.