Justin CrooksModeratorJustin Crooks July 15, 2020 at 3:53 pm in reply to: FTDI Incompatabilities? //php bbp_reply_id(); ?>
It’s a Cypress FX3 chip, CYUSB3014
Justin CrooksModeratorJustin Crooks July 15, 2020 at 9:40 am in reply to: FTDI Incompatabilities? //php bbp_reply_id(); ?>
Kaiser,
I see two possibilities here.
1) The sheer data rate of the SM200B is causing problems, or
2) When you installed the software, you also installed the FTDI driver for our SA44B, which is somehow incompatible with the VN-310.If you re-install the VN-310 driver does it help? If you use FTDI’s “CDM uninstaller” with the VID/PID of the VN-310, and then plug it in, does it help?
Justin CrooksModeratorJustin Crooks June 30, 2020 at 9:46 am in reply to: Spectrum Analyser Question //php bbp_reply_id(); ?>
Matthew,
I have never seen that before, but if you can write a little code, parse the wav file, probably window the data, do an FFT, threshold detect (1 for above threshold, 0 for below threshold), invert the threshold array, inverse FFT, then scale. I am not aware of an application for this, but if you have one, something like this would be a good way forward.
Justin CrooksModeratorJustin Crooks June 11, 2020 at 9:36 am in reply to: Generating Gaussian noise with VSG25A //php bbp_reply_id(); ?>
Geev,
The I/Q pattern memory is pretty short (2k samples), but yes, a short Gaussian waveform is possible. You can also look at 1001 tones with random phase relationship, which will have many of the same properties as bandwidth-limited Gaussian noise.For a better Gaussian white noise generator, consider the VSG60A.
Justin CrooksModeratorJustin Crooks May 4, 2020 at 9:54 am in reply to: How to record series of spectrums //php bbp_reply_id(); ?>
Kefei,
There are a few ways to do this. You could record the sweeps in Spike (you may need an SM200 in real-time mode, span >160 MHz, for 1 ms sweeps).
For more control, you could simply record 1 second of I/Q and do your own windowed FFTs. With the SM200B, this would give you up to 2 seconds at 160 MHz span.
A third option is the waterfall spectrum plot in zero span mode. In this mode, the biggest limitation is probably the 1 second requirement. This would set a pretty low I/Q sample rate, since you are limited to a few million I/Q data points.
Justin CrooksModeratorVolker,
Yes, this is something we have discussed. It is not something we have scheduled at this point, but it is definitely something we hope to add in the future. It would probably not be one of our existing tracking generators, but something geared towards the frequency range and performance of the SM series.
Justin CrooksModerator- This reply was modified 5 years, 1 month ago by
Justin Crooks.
- This reply was modified 5 years, 1 month ago by
Justin Crooks.
Justin Crooks March 25, 2020 at 1:18 pm in reply to: Non-filtered BPSK on VSG25A has weird spectrum //php bbp_reply_id(); ?>
Gary,
There are a few things going on here. It sounds like you are aware there is no baseband reconstruction filter on the VSG25A. But we also apply an inverse sinc correction to the in-band data to flatten it.
PN9 is a PRBS9 sequence (linear feedback shift register) that basically repeats every 511 symbols. We upsample each symbol by a factor of 4 to produce 2044 I/Q samples, which get transmitted in a loop.
Even with no raised cosine filter, the upsample process will result in gaps in the sin(x)/x curve.
Hopefully this helps.
Justin CrooksModeratorJustin Crooks March 17, 2020 at 9:34 am in reply to: Suitable preamp for 10X scope probes into SA44B //php bbp_reply_id(); ?>
If it helps there are schematics online. Typically a 1 Mohm input impedance to a non-inverting unity gain op amp, with a 50 ohm series resistor (or resistor to bring it to 50 ohms) on the output. You may want to add a ceramic [0.1 – 10] uF DC blocking cap, depending on your low frequency coverage needs.
Justin CrooksModeratorJustin Crooks March 16, 2020 at 3:14 pm in reply to: Suitable preamp for 10X scope probes into SA44B //php bbp_reply_id(); ?>
We have a probe that works from about 100 kHz to 4 GHz, with about a 500 ohm AC load to your circuit (DC blocked). We are not set up for a FET probe, which would be what you would need for a high impedance load to your circuit but a 50 ohm output impedance. This is more of an oscilloscope function.
Justin CrooksModeratorJustin Crooks February 19, 2020 at 9:40 am in reply to: How to read IQ data I matlab //php bbp_reply_id(); ?>
Kefei,
To convert I/Q data to spectrum, you will need to perform overlapping windowed FFTs. Typically what this looks like, if you are doing a 1024-point complex FFT, is every N samples (for your application this would be every 1 ms), you will take the last 1024 I/Q samples, window them (multiply by a window function, like a flat top or hamming window), and then FFT them to generate a spectrum.
Justin CrooksModeratorWe have not interfaced SDR sharp. One approach would be for your friend to develop a plugin, to connect the BB60C to SDR# directly. Another would be to use our API to save I/Q data as a WAV file and try to play that in SDR#.
Justin CrooksModeratorJustin Crooks January 15, 2020 at 4:06 pm in reply to: Another Synchronization Question //php bbp_reply_id(); ?>
Basically, queued fast sweeps just go. Unfortunately there is a little bit of uncertainty each time the LO changes. We check to see if it has settled, and add a few microseconds delay if it has not, so even if they started in sync it would not last.
We could add this to our list of feature requests. Most likely this would be a feature we would add to a future product (multi-channel receiver) due to the complexities of synchronizing two standalone devices.
If you could scan your spectrum more slowly, 40 MHz at a time, you could use I/Q streaming and a shared external trigger to get them sync’d within a handful of nanoseconds.
Justin CrooksModeratorJustin Crooks January 2, 2020 at 11:50 am in reply to: IP2 & IP3 CALCULATION IN SPIKE //php bbp_reply_id(); ?>
Keysight has a good IP3 guide: https://www.keysight.com/upload/cmc_upload/All/ENA_IMD_Measurement_Summary.pdf
Justin CrooksModeratorJustin Crooks November 27, 2019 at 11:37 am in reply to: GPS disciplined 10MHz ref clock for SA44B //php bbp_reply_id(); ?>
That model should work just fine. A 10 MHz square wave will provide slightly better performance, but it looks like the output is +10 dBm or higher, which is good.
Justin CrooksModerator- This reply was modified 5 years, 5 months ago by
Justin Crooks.
Justin Crooks November 13, 2019 at 10:31 am in reply to: Different performance on similar PC //php bbp_reply_id(); ?>
For what it’s worth, I get 375 ms on my PC for the same sweep. Unfortunately, sweep time, especially on narrow spans, is not deterministic. There are settings applied, and then acknowledged, several times per sweep. So the timing will vary based on your USB read/write latency, which can vary for a lot of reasons. Spans of several MHz will generally be more consistent since the sweep is controlled more in firmware than software.
Justin CrooksModeratorJustin Crooks November 11, 2019 at 8:48 am in reply to: Different performance on similar PC //php bbp_reply_id(); ?>
Do you mind if I ask what the center frequency, span, RBW and VBW settings are? What is the model of the i7 processors? Is spur reject on for both cases?
It is possible for a sweep to take twice as long if the PC is significantly slower, and smaller variations are fairly common, but sweeping twice as fast on a similar PC seems unusual.
Justin CrooksModeratorJustin Crooks October 10, 2019 at 3:46 pm in reply to: SA44B change amplitude with Ref Lev //php bbp_reply_id(); ?>
Around 2016 we started adding a resistor to provide a weak return path to ground because some units would develop this condition you described over time. I think the problem stems from the fact that there’s no return path to ground when the attenuator is 0 dB, and a tiny leakage current from the attenuator can essentially blind the RF switch until a ground path exists.
If you can set the attenuator to 5 dB, or add a 1 dB external attenuator, this problem should clear up. Otherwise, you can get an RMA, send it in, and we can add the resistor.
Justin CrooksModeratorCristin,
There are several factors that go in to why these measurements are different. These include the software image rejection algorithm, the noise floor / sensitivity of the SA44B in the selected mode, and the effective video bandwidth and averaging method based on detector settings.
If your goal is noise spectral density measurements, the preferred instrument is the BB60C or SM200A. This is because the SA44B does not have hardware-based image rejection, so there is noise present from both sidebands (even if the signal from one sideband is rejected in software).
That being said, real-time mode, power average is a good way of measuring the noise energy (sum of energy at desired frequency and image frequency–subtract 3 dB typically for noise at desired frequency only).
For a lower noise floor, using sweep mode (image rejection ON), power, average, VBW <= 1/10 RBW, will use a more sensitive detector circuit and provide a lower noise floor than real-time mode.
Justin CrooksModerator- This reply was modified 5 years, 6 months ago by
Justin Crooks.
Justin Crooks October 7, 2019 at 10:26 am in reply to: Power supply DC connector //php bbp_reply_id(); ?>
Miran,
One option for powering down the SM200B is to shut down the PC controlling it. Most PCs allow you to select whether or not the USB is powered when it is shut down. If the USB 5V is turned off, the SM200B will turn off completely (I believe it is something like a milliwatt of power draw).
If a mechanical switch is what you want, any switch that is rated for at least 20V, 5A should be sufficient.
Justin CrooksModeratorJustin Crooks September 25, 2019 at 9:58 am in reply to: Rx Converter measuring //php bbp_reply_id(); ?>
If an approximate conversion loss (+/- 3 dB) is acceptable, you can sweep the output range with the SA44B, and then step across the input range with the TG manual controls, using MAX HOLD on a trace.
For better accuracy, an RF power meter could be used to measure the actual TG output power at each frequency and you could subtract this from the SA44B’s reading.