Forums › SP Series Discussions › Strange spurs in SP145 without input signal
- This topic has 1 reply, 2 voices, and was last updated 3 months, 1 week ago by
Justin Crooks.
- AuthorPosts
MehdiParticipantHello. I did a simple test to compare SP145’s noise floor to BB60D, and observed a strange behaviour in SP145. Please check the screenshots.
Is this related to the fact that SP145 doesn’t have a preselector filter bank?
Settings are equal:
30MHZ-6GHZ
10KHZ RBW and 3KHZ VBW
Trace mode Average (100 times averaging)
Ref level -30dBm
Attenuation 0dB
Input port terminated with a high quality 50 ohm load.
Both tested on the same version of Spike on the same laptop in the same environment.Bests
MehdiAttachments:
You must be logged in to view attached files.
Justin CrooksModerator- This reply was modified 3 months, 1 week ago by
Justin Crooks.
Mehdi,
Signals present when the input is terminated are “residual responses” or sometimes “residual spurious”.
The specification for the BB60D is -120 dBm. This is one of the strengths of the BB60D.
The specification for the SP145 is -103 dBm, similar to the BB60C.
This is not directly related to the preselector, and is more closely related to filtering and isolation differences in the architectures.- This reply was modified 3 months, 1 week ago by