Forums › SP Series Discussions › SP145 wrong amplitude reading and IF overload
- This topic has 3 replies, 2 voices, and was last updated 3 months, 1 week ago by
Andrew.
- AuthorPosts
MehdiParticipantHi. I was doing a simple test and comparison between SP145 and BB60D where I encountered a strange issue:
I generate a -5dBm 2GHZ CW signal using VSG60A and connect it to BB60D and SP145, to show their difference in phase noise, in narrow-band settings, but SP145 shows IF overload and also wrong amplitude, while BB60D shows everything properly.
Please check the screenshots.
Settings on Spike:
Center freq: 2GHZ
Span: 50KHZ
RBW: 100HZ
VBW: 10HZ
Trace mode: Average (50 averages)
Att: 0dB
Ref level: 0dBmAttachments:
You must be logged in to view attached files.
AndrewModeratorMehdi,
You specify an attenuation of 0dB, which for the SP145 I believe is overriding your reference level control. If you want reference level to control the sensitivity of the receiver we recommend leaving all gain/atten/preamp values to their defaults. The reference level control alone will properly set these controls behind the scenes for the best dynamic range. If you change attenuation back to Auto, with a ref level of 0dBm, your measurement should work again on the SP145.
The reason you probably don’t see the same issue on the BB60D is because you have to set both gain and atten before it override the Ref level. Again, we still recommend leaving everything auto and just using ref level, unless you have a specific reason not to.
MehdiParticipantThanks. Setting SP145 to auto indeed fixes this.
I wanted to get the absolute minimum noise (not having any attenuation or amplifier in the input path) but this is fine for my use case.
AndrewModeratorA reference level of -20dBm on the SP145 achieves the minimum noise floor, but will limit input signals to ~-20dBm before IF overload occurs. On the BB60D, the minimum noise floor is achieved at -30dBm reference level.
- AuthorPosts
You must be logged in to reply to this topic.